8 ms·
Hey, lowly Bio undergrad here, but how are you able to skip the RNA extraction step? I read the paper and you use the viral transport medium, but wouldn't you h
by abhisuri97 7y ago
Hey, lowly Bio undergrad here, but how are you able to skip the RNA extraction step? I read the paper and you use the viral transport medium, but wouldn't you have to also purify RNA from that (or is it just much easier to extract RNA from that medium)? I also dived into the paper behind the "skip the RNA extraction step" methodology and it basically seems to swap out one RNA extraction kit for another (Qiagen RNeasy Mini kit and the Qiagen RNeasy Micro kit). Couldn't shifting kits from one provider to another introduce supply chain strain? (or am I just oversimplifying it?)
- billiontoone 7y agoWe have the data both with extraction and without and show that it does not make a difference with qSanger. In Figure 4, we add VTM directly to PCR reactions. Seracare VTM samples has SARS-CoV-2 viral RNA in a different capsid to prevent infections in a research setting, but otherwise it reflects real-world VTM samples (and much more realistic than even what EUA requires). By the way, this robustness is completely expected, as any impurities in VTM would impact spike-in and endogenous viral amplification equally for end-point PCR (so their ratio stays the. same). This is not necessarily true for qPCR where an impurity (caused by lack of RNA extraction) can potentially cause a positive sample look like a negative when the viral RNA does not RT-PCR.
- dvdt 7y agoThanks for the question! The goal of skipping RNA extraction is to decrease the amount of labor necessary for processing samples and also to eliminate a dependency on RNA extraction reagents that have recently become difficult to find. The FDA is very strict about the specific brand and model of kit you use, so showing that you can swap out one RNA kit for another is actually very useful because you will have alleviated some of the supply chain strain (although I agree at high enough load both supply chains will then become limiting). The way currently available COVID-19 testing works is by detection of viral RNA. Since the amount of viral RNA in a patient sample is too low to detect directly, we first need to amplify it by PCR. However, this viral RNA is packaged within all sorts of proteins and lipids that could make it inaccessible to amplification unless they are first purified away. Furthermore, the sample is shipped in "viral transport medium", which is essentially a cocktail of chemicals designed to preserve the virus. Unfortunately, these preservatives often have the side effect of interfering with PCR amplification, so these too need to be purified from the sample. However, since RNA extraction is usually the most laborious part of the assay, there has been a lot of interest in optimizing the amplification so that it is resilient to all of these impurities. The preprint referenced in our manuscript (https://www.biorxiv.org/content/10.1101/2020.03.20.001008v1 https://www.biorxiv.org/content/10.1101/2020.03.20.001008v1) gave us the initial idea that this could be possible, and much of it comes down to the choice of amplification method (e.g. choice of enzymes and buffers) that you choose. However, even when you choose a "good" enzyme and buffer, you will still suffer an amplification penalty, and this will cause you to return a false-negative on some affected samples because there was so little virus in the sample to begin with. The innovation we have is to spike-in a correspondingly low level of DNA to the reaction mixture. That way, if you see the low level of DNA without seeing any viral signal, you can be assured that the amplification still worked and that there truly is no virus in the sample.
- abhisuri97 7y agoThank you so much for that explanation!
- treis 7y ago>the amount of viral RNA in a patient sample is too low to detect directly, How many tests have you run on patient samples?
- mattmanser 7y agoIn the UK they're saying there's a shortage of swabs and pipettes even, do you not need these too? Also, in the UK our independent and uni labs have been saying for almost a week they could extract the RNA differently but the NHS have a fixed approved way that they won't change. - Are the chemicals you're using more common or would there just be a new shortage of different chemicals? - Is there a risk you'd be creating a test that didn't work very well, and the US would end up with a bunch of useless tests (e.g. Italy had to abandon a bunch of Chinese sourced tests, UK's anti-body tests are ineffective)?
- dvdt 7y agoOur technique would still be affected by shortages in specimen collection (like swabs). Purely speculative, but I think if swabs remain an issue for too long, alternatives could start coming online, such as even using qtips + saline (no idea if it works, it's just an example). The current swab + Universal / Viral Transport Medium combo is optimized for flexibility; it is designed to work across a very broad range of viruses and bacteria that have different viral loads and shedding characteristics. The current pandemic is pretty much COVID-19 only, so I think it's a priori feasible that a specimen collection procedure can be found that uses common materials. We did try early on to see if saline or other buffers affected the performance of the assay, and it worked fine in those conditions. We use fairly standard chemicals. I haven't heard from our suppliers about shortages for the chemicals we use. Chemicals and enzymes tend to be relatively fast to scale up for bulk manufacturing. There's always manufacturing risk that a product will not work as expected. In fact, the first COVID-19 test developed by the CDC did not work as expected, and this delayed testing by several weeks. We de-risk this as much as possible by performing experiments as early as possible, akin to the fail fast mentality of checking for the highest risk failure modes first. Since we don't have a national healthcare system in the US, the manufacturer takes on the vast majority of the risk of a defective product.
- mjul 7y agoThe RNA extraction can also be done without reagents by using a heat reaction on the sample similar to boiling an egg. It is a 5-minute process. According to an older scientist, Anders Fomsgaard at the Danish Serum Institute, this is how they did it ”in the old days”. He is the father of one of the authors. This eliminates supply chain problems for reagents and was shared quickly to help in Spain. Preprint by Fomsgaard and Rosenstierne: https://www.medrxiv.org/content/10.1101/2020.03.27.20044495v1 https://www.medrxiv.org/content/10.1101/2020.03.27.20044495v...
- raspasov 7y agoThis is amazing.